DS213 gain calibration

[复制链接]

2

主题

17

回帖

236

积分

中级会员

积分
236
发表于 2019-7-1 19:08:48 | 显示全部楼层 |阅读模式
Hi.

How do you calibrate the gain of a channel?

The K4+K3 combination calibrates the ground level.
Rotating the dial A while "CAL" option is selected will only shift the entire range, but doesn't change the gain.
Effectively, it is the same as ground level shifting.

The voltage measured are alwas off by some fixed value, and if I try to compensate it by rotating A dial with "CAL" option, I would also shift the zero.

What's the correct method of scope calibration using external voltage sources?


回复

使用道具 举报

197

主题

1357

回帖

3052

积分

管理员

积分
3052

社区居民忠实会员

发表于 2019-7-2 15:32:37 | 显示全部楼层
You use DS203 or DS213?
DS213 is auto calibrate, no need choice "CAL"
回复

使用道具 举报

2

主题

17

回帖

236

积分

中级会员

积分
236
 楼主| 发表于 2019-7-2 16:47:26 | 显示全部楼层
It's DS213.

It does something when calibrating, but what's the procedure?
Shorting the probe? Connecting to own signal generator output?
Any ways to use external precise voltage source?

In any case, the measured values are always somewhat off from the known source voltage.
And performing auto calibration without correct procedure makes the next measurements even worse
回复

使用道具 举报

197

主题

1357

回帖

3052

积分

管理员

积分
3052

社区居民忠实会员

发表于 2019-7-5 09:28:12 | 显示全部楼层
The channel gain of the DS213 is fixed.
No calibration is required.
‘CAL’ is used when auto-calibration fails in extreme cases and is only used for zero calibration.
回复

使用道具 举报

2

主题

17

回帖

236

积分

中级会员

积分
236
 楼主| 发表于 2019-7-5 17:21:51 | 显示全部楼层
Well, that's not really good.

My multimeter and lab PSU agree pretty well on the voltages. Say, 3.3 v selected. The DS213 measures 3.08v.
For 2.5v source DS213 measures 2.34v, and for 1.8v source DS213 measures 1.68v.
The observed scale is pretty much the same, ~0.9333

It's understandable that scaling the values for the waveform takes precious clock cycles, but at least the summary values shown in the bottom row (Vmax, Vmin, Vp-p,  /etc) could be scaled. They don't update very frequently, so scaling it wouldn't hurt the performance.
回复

使用道具 举报

197

主题

1357

回帖

3052

积分

管理员

积分
3052

社区居民忠实会员

发表于 2019-7-8 10:40:49 | 显示全部楼层
The next version will add a new feature (gain calibration for users to accurately measure voltage)
In addition, oscilloscope sampling is affected by input signal ripple, probe matching, etc. It is not recommended for accurate voltage measurement.
回复

使用道具 举报

1

主题

6

回帖

12

积分

新手上路

积分
12
发表于 2023-4-4 08:05:38 | 显示全部楼层
Where is this new version? Not whats there now, I mean newer! This thing is being beat by a $40 scope on amazon. I know this beast has some potential but you guys seem to have given up on it, although you are still selling them by the truck loads!
回复

使用道具 举报

197

主题

1357

回帖

3052

积分

管理员

积分
3052

社区居民忠实会员

发表于 2023-5-24 09:11:02 | 显示全部楼层
Sorry, no timeline to make public。
回复

使用道具 举报

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

Archiver|手机版|MINIWARE产品技术交流 迷你工具-智能烙铁-加热平台-示波器-体感电动螺丝刀-数字电源-智能镊子 ( 粤ICP备07030012号-1 )

GMT+8, 2024-11-21 23:24 , Processed in 0.153532 second(s), 21 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表